Re: Lipofectamine auto fluorescence

From: Simon_Q_Rice@sbphrd.com
Date: Thu Sep 06 2001 - 10:16:27 EST


Tami,
Are you sure your autofluorescence is caused by the lipofectamine?  Don't
you normally wash off the DNA/lipofectamine complex at least 24 hours
before analysis?  If not I suggest you do and replace with fresh medium,
then leave your cells at least 24/48 hours before analysis.  The other
thing you might try is resuspending your cells in a phenol-red free medium
just prior to analysis.  Phenol red can cause autofluorescence.

I hope that's helpful.

Kind regards

Simon




trosario@cellmate.cb.uga.edu on 31-Aug-2001 21:52



To: cyto-inbox
cc:
Subject:  Lipofectamine auto fluorescence



We are transfecting COS 7 cells with a plasmid containing GFP using
lipofectamine.  We see a large amount of auto fluorescence from the
lipofectamine alone in our controls.   Has anyone else experienced
this?  If so, how can we remedy the situation.
Thanks for your help,
Tami Rosario
Department of Cellular Biology
University of Georgia