apoptosis in adherent cells (and tissues?)

From: Richard Haugland (richard.haugland@probes.com)
Date: Fri Aug 03 2001 - 23:29:35 EST


I recommend trying our (Molecular Probe's) YO-PRO-1 to detect apoptotic
cells in an adherent-cell population (or tissues?). The reagent is a
nucleic acid stain that is nonfluorescent except when bound to
intracellular nucleic acids. It is selectively permeant to membranes of
apoptotic cells and has been used for this purpose several times in flow
cytometry and imaging. Since it is also permeant to dead cells, I
recommend combining it with propidium iodide to detect any dead cells
(which will have bright red fluorescence).  Most likely both would be
used at about 1 uM or less in standard media (minus phenol red) for
perhaps 15 minutes, using a fluorescein long-pass filter to discriminate
the staining of apoptotic and dead cells (live cells should be unstained
but could likely be stained blue fluorescent with Hoechst 33342).  These
dyes will probably NOT require any wash step after staining if used
dilute enough.  Ethidium homodimer-1 at an even lower concentration
would definitely not require a wash step to detect the dead-cell
population.

The flow cytometric version of the assay is described (in the Product
Information Sheet) here:

http://www.probes.com/servlets/product?region=Select&item=13243

(although I see that the abscissa in the figure in the Product
Information Sheet that is now marked "FITC" fluorescence" really would
be better marked "YO-PRO-1 fluorescence, although FITC and YO-PRO-1 use
the same optical filters and both are FL 1 measurements in a flow
cytometry).

I would be interested in hearing form anyone who has tried this in
imaging of adherent, partially apoptotic cells and we plan to try the
3-color experiment soon.

Some Pubmed links on doing this (by microscopy):

http://www.ncbi.nlm.nih.gov/entrez/query.fcgi?cmd=Retrieve&db=PubMed&list_uids=9749586&dopt=Abstract

http://www.ncbi.nlm.nih.gov/entrez/query.fcgi?cmd=Retrieve&db=PubMed&list_uids=7561136&dopt=Abstract

http://www.ncbi.nlm.nih.gov/entrez/query.fcgi?cmd=Retrieve&db=PubMed&list_uids=10870639&dopt=Abstract



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:27 EST