With interest we saw the results of your study on counting of CD4, CD8 and
CD3 using a bead based single platform device.I donīt think that your
underestimation has anything to do with the anticoagulants that you used in
your study. We found the same 10 - 20% underestimation for such subclasses
by comparing the results of a volumetrical flow cytometer measurement with
those of a dual platform system. In case of CD34 cells, the loss was much
more dramatic. Depending on the lysing reagent used, the underestimation
was in the range of 34 - 45%.
In a detailed study we recognized the lysing reagents as the reason for
this cell loss, by membrane destruction. Using the dual platform mode,
these effects are compensated by the hematology counter results in
connection with the CD45 numbers.
We found that the destruction of some leucocytes by different lysing
reagents is well known in the literature, but obviously not enough being
taken into consideration in the companies supported protocols.
The best solution could be a non lyse protocol, we found one which
obviously works well.