BD Calibur threshold behavior

From: Marty Bigos (mbigos@gladstone.ucsf.edu)
Date: Thu Jul 19 2001 - 18:49:43 EST


I've come across this odd threshold behavior on the BD FACSCalibur
which I thought I would mention since I haven't seen any notice of it
on this list. It can be summarized as follows:

The measurement values of first laser fluorescent signals decreases
if a forward scatter threshold is set "too near" the forward scatter
signal of the particle.

Since it is not uncommon to threshold high to eliminate debris, I
looked at this a little closer (data below). Although I don't think
this will affect the majority of experiments, there could be cases
where it does. BD is aware of this behavior (PIR submitted through
Dave Houck). The exact amount of measurement change varies from
machine to machine, so the data below is more qualitative than
quantitative.

In the "bead compare" document, the left column of plots are FSc x
SSc with different FSc thresholds indicated. The right column are the
single beads with the median listed. For this set, the bead
measurement dropped from 619 units to 284 units.

The "Bright cell stain" document shows this affect on the a bright
positive population (CD45) of PBLs. The left column shows FSc
histograms at various threshold levels. The right column is the CD45
positive lymphs, showing a decrease in median from 1537 to 1426. If
one were measuring percent positive populations, this should not
affect the results.

The "Dim cell stain" shows this artifact on dimly staining cells
(IF-gamma). The columns are as above. There is about a 5% decrease in
the percent positive - in this case not a significant change.
However, for other instruments, this may be greater (or less).

Caveat emptor!




-- Marty Gladstone Institutes Flow Core mbigos@gladstone.ucsf.edu 695-3832



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:26 EST