Bob Zucker wrote-

"On a recent advertisement ( post card)  by MoFlo they implied that
digital processing of data (DIVA) was not as good as the MO FLo
type
of analysis. I thought that digital processing of Coulter and BD was
superior to the older designed electronics. Although the
distributions
from MoFlo  do indeed look superior to	a BD Vantage ( postcard),
the
Cytomation people are implying that digital electronics are inferior
to their current circuitry. Any opinions on this claim of the new BD
equipment with digital electronics as being inferior. Now I am
confused as I thought digital was the new and improved way to do
flow
cytometry. Is this just an advertisement  from a company that does
not
want to get on the digital bandwagon or is it because they do not have
it and don't want to redesign their equipment with it just yet.  I
don't want to get into a debate on the virtues of the different
manufacturers but it is important to clarify this point for the flow
cytometry community."

For the record, I do consult for Cytomation and Luminex, and have
consulted for Beckman Coulter and B-D.	And I have certainly been as
vocal an advocate of digital processing in flow cytometry as anybody.

The short answer to Bob's question is that, while, in principle, it is
preferable to do as much of the signal processing as possible
digitally (i.e., digital thresholding, baseline restoration,
computation of pulse peak height, area, and width), *none* of the
existing electronics achieve ideal performance in practice.  Different
companies have made different tradeoffs.

Looking at Cytomation's post card ad, what was obvious was that the
data from the blank bead in an 8-peak bead set measured using B-D's
DiVa electronics were somewhat "granular" - Cytomation described the
peak as a "picket fence".  This effect is also notable in the bottom
half of the bottom decade on digital electronics I have built (see
Shapiro et al, Cytometry 33:280-287, 1998, particularly Fig. 1 and
associated text on p. 285).  It is primarily an artifact of converter
resolution.  The "picket fence" does not appear on the distribution
collected using a Cytomation MoFlo.  Is this necessarily better?

Well, to tell the truth, if I had to pick a feature of the
distributions to be impressed by, it would be the distance between the
blank peak and the next highest intensity peak, which was apparently
greater in the sample collected by the MoFlo than in the sample
collected by the instrument with DiVa electronics, suggesting  that
the MoFlo had higher sensitivity, which would depend more on optics
and detectors than on electronics.  However, one ought never to form
an impression of comparative performance from a single sample
comparison.  B-D hasn't sent out its post cards yet.

In terms of the actual performance of the electronics, Mario Roederer,
who has had extensive experience with B-D optics using Cytomation
electronics, and is now using a B-D system with DiVa electronics,
reported (also on this Mailing List):

"In our preliminary evaluation of the digital electronics, we get as
good or better resolution, separation, and sensitivity using the
digital signals as compared to the analog signals on the same bench
(i.e., DiVa vs. SE). In a preliminary comparison of some 8-color
samples with those analyzed at our original machine at Stanford (which
uses the Cytomation electronics), performance was quite similar."

The primary problem that faces the users and developers of cytometry
apparatus is this:  our data values frequently span a large range -
generally not quite the four decades we take as standard, but
certainly close in many situations, particularly immunofluorescence
analysis.

Immunofluorescence analysis gives rise to a second major problem; the
emission spectra of the probes or labels used overlap, necessitating
that hardware and/or software compensation be used to produce correct
values for the fluorescence of individual labels from the values of
fluorescence in particular spectral bands.

In order to display weak and much stronger fluorescence signals on the
same scale, it has become standard practice to convert from linear to
logarithmic values/scales.  Historically, this was first done using
hardware, i.e., analog logarithmic amplifiers, familiarly (and
familiarity has bred contempt) known as log amps.  Now, most of us
have known since high school that you can look up the logarithm of any
number greater than zero in a table and rely on the answer being as
accurate and precise as the table.  Unfortunately, while, in
principle, a log amp takes a voltage (or current) as its input and
delivers an output proportional to the logarithm of its input, log
amps are not usually as good as log tables.  They are typically
specified as being accurate within 1/2 decibel; while this figure
sounds pretty good when you're stereo shopping, it turns out to mean
plus-or-minus 6 per cent.  Most log amps get worse at the low end;
some do at the high end.

Now, facing the facts, +/- 6% is probably as good as you need for
immunofluorescence analysis; about the only place in cytometry you
want more precision is in nuclear or chromosomal DNA analysis, and we
all do that on a linear scale, eliminating any imprecision associated
with the log conversion.  However, many log amps change their response
characteristics on a day-to-day basis due to temperature fluctuations,
etc., so it might make life easier to do without them.	It definitely
makes life easier to do without them when you use four or more colors
of fluorescence, because compensation has to be done on linear
signals, and the typical approach to doing this in an instrument with
log amps involves putting the compensation circuit between the
detector preamplifier outputs and the log amp inputs, and the
complexity of the compensation circuit goes up faster than the number
of colors used.

The way to eliminate log amps is to collect linear data in a
sufficiently precise manner to allow you to use a good old-fashioned
log table to go from a linear to a logarithmic scale. This requires
the use of analog-to-digital converters (ADC's) Until recently, most
flow cytometers used 8-bit (256-channel) or 10-bit (1024 channel)
ADC's.	However, to accurately capture data with a 4 decade dynamic
range, where the weakest signal is 1/10,000 the intensity of the
strongest, one needs a higher resolution ADC.  A 14-bit ADC has 16,382
channels, and will theoretically respond, but an ADC with 16 (65,536
channels) or more bits is preferable.

Flow cytometers used to have 8- and 10-bit ADC's because higher
resolution ADC's either didn't exist or were prohibitively expensive.
By the late 1980's or early 1990's, 16-bit ADC's which could convert
data in a few microseconds became available.  This led to my
development of the digital electronics described in the Cytometry
article cited above and to Coulter's development of the electronics in
the Epics XL.  Neither of these instruments did digital pulse
processing; instead, they took analog signals representing held values
of the peak and/or integral of a pulse and produced a single
high-resolution value.

My electronics use a 16-bit converter to produce linear values between
0 and 65,535; the lowest linear value in the lowest decade on the
four-decade scale is 7, and the lowest values of the higher decades
are 64, 656, and 6554.	I can convert from these 16-bit linear values
to a 256-channel (8-bit) logarithmic scale, but, at the low end of the
lowest decade, one value on the linear scale corresponds to multiple
values on the log scale, and using only a single value generates the
"picket fence" appearance shown in the figure in my paper.

The XL electronics use a raw signal and an amplified signal, with a
nominal 15-bit converter, to generate 20-bit linear data.  This allows
conversion from 20-bit linear values to a 1,024-channel (10-bit) log
scale, without the "picket fence" effect.

Software packages which provide the facility for digital compensation
must convert log data to linear data so compensation can be applied.
Data taken from log amps via analog peak detectors or integrators,
converted to only 10-bit (1,024-channel) precision, then converted to
a linear scale, compensated, and put back on a log scale, show a
pronounced "picket fence" effect; software that operates on such data
usually adds random numbers to reduce or eliminate the "picket fence"
effect.  This makes the data look better, but cannot help but decrease
accuracy.

High-resolution digitization of signals held in peak detectors or
integrators can be accomplished using ADC's with conversion times of 1
microsecond or more, equivalent to continuous conversion rates of less
than 1 MHz. Digital pulse processing eliminates the need for, and any
inaccuracies or limitations of, analog peak detectors or integrators
(or pulse width measurement circuits), but requires multiple samples,
or "slices" of a pulse, and thus demands faster ADC's.

My colleagues and I have calculated that 8 slices of a pulse can yield
an integral (area) nearly as accurate as can be obtained using 16
slices, and that 32 slices are not significantly better than 16.
However, peak and width measurements get notably worse as the number
of slices drops, as you might expect, because the relevant slices are
likely to be farther away from the peak or threshold values.

The earliest digital pulse processing in flow cytometry was done by
Leon Wheeless and his colleagues in the 1970's, in "slit-scanning"
apparatus, with relatively slow flow rates and long pulse durations.
They could operate 8-bit ADC's at frequencies of several MHz,
producing dozens or even hundreds of points per pulse.	Digital pulse
processing was incorporated into other instruments, but did not use
ADC's with high enough resolution to provide the 4-decade dynamic
range required for immunofluorescence work.

A 12-bit converter has 4,096 channels, and produces output values
between 0 and 4,095.  If you calculated a pulse integral by taking 8
slices of a pulse with a 12-bit ADC and summing the values, you would
think you'd be able to cover a range between 0 and 32,760, which
should encompass a 4-decade dynamic range.  However, the actual range
achievable is smaller, for two reasons.  First, the signal at the
lowest end of the range is 1 part in 10,000, and, to get the ADC to
put out an output of 1 instead of 0, you need about 2.5 times that, or
1 part in 4,096, so the 12-bit ADC won't respond to signals near the
bottom end of the range.  Second, the highest possible value, 32, 760,
crops up only when the input pulse is flat-topped, that is when the
signal is "clipped".  The maximum value for the integral that is
accurate is that which would be obtained from a pulse in which only
the peak (1 or at most 2 slices) reached the value 4,095, and, for a
Gaussian pulse, this is around 20,000.

My calculations show that, no matter whether 8, 16, or 32 slices are
taken, digital pulse integration using a 14-bit converter gives
substantial inaccuracy (close to 50 per cent) at the bottom of the
range (bottom half of the bottom decade); however, the situation may
not be any better when you use log amps because they are also almost
certainly inaccurate at the low end.  B-D apparently agrees with all
this.  The DiVa electronics use 14-bit ADC's operating at a frequency
of 10 MHz; when I have seen them in operation, they were taking 32
slices of a 3.2 microsecond pulse.  If the pulse were shorter, the
integral would have to be computed from a smaller number of slices
(e.g., for the 800 nsec pulse duration typical in a Cytomation MoFlo,
it would only be possible to take 8 slices).  At the low end of the
lowest decade, the DiVa electronics come up with the "picket fence"
effect.

However, the low end is where the blanks and the noise are; the
signals of greatest interest are higher up on the scale, and either
digital pulse processing or high-resolution digitization of held peaks
or integrals should be more accurate there.  In my system, accuracy
seems to be within +/- 2%; as mentioned above, accuracy  with a log
amp somewhere in the loop is unlikely to be better than +/- 6%.

Cytomation does have a log amp in the loop; they do digital
compensation by performing a 16-bit conversion of held peaks or
integrals of the log amplified signal, converting to linear for the
calculation, and then converting back to log.  This largely avoids the
"picket fence" effect at the low end, but probably yields lower
accuracy further up the scale than would be expected from a full
digital system without a log amp.  However, Mario Roederer's comments
quoted above suggest that it's hard to tell the difference.

For now, I have to point out that neither B-D nor Cytomation has
achieved an ideal solution, which would be digital pulse processing
with at least 16 slices from an 18- or 20- bit converter (even a
16-bit converter is about 14% inaccurate at the low end when doing
digital pulse integration).  And, right now, there aren't any 18- or
20-bit converters fast enough to do the job.  There are 16-bit, 5- and
maybe 10-MHz parts, and there are now 14-bit, 100 MHz ADC's, but, so
far, nothing with higher resolution at even 2.5 MHz.

There are, however, 24-bit converters which sample as fast as 192 kHz,
which are relatively inexpensive by virtue of being used for digital
audio recording.  Using these to digitize held analog peaks or
integrals could provide higher accuracy and precision than are now
available, and keep up with the data rates required for high speed
sorting.  This sounds good, but, to make it work, you need analog peak
detectors and/or integrators good to 1 part in 10,000, and these are,
at best, extremely difficult to come by.  The Beckman Coulter XL has
integrators good to 1 part in 10,000; the splitting and amplification
of the input signal to the ADC's was necessary when the instrument was
designed, years ago, because a 20-bit converter was unavailable.
Whether the XL electronics would work in a system running 100,000
cells/sec, I don't know.  It gets easier to cover a large dynamic
range if you split signals into several intensity levels.  My
colleagues and I are looking into these approaches, and encourage
others to do the same.

Note that progress in cytometry electronics depends on our industry,
which consumes economically insignificant quantities of components,
being able to use parts developed for other industries - notably the
entertainment industry.  If digital audio, digital video, and games
will benefit from higher-speed, higher resolution converters, they'll
become available to us; if not, we'll have to continue to compromise.

Bottom line: If you are buying a high-speed sorter, whether or not it
has full digital electronics is probably not the most important single
factor to consider in your decision making, at least for now.  Ads are
ads; what you need to do is get demos run on the samples of interest
to you.  At today's prices, the manufacturers can afford to work for
their money.

If you want to read more on this subject, it will be covered in great
detail in the 4th Edition of Practical Flow Cytometry, due out at next
May's ISAC meeting.  Please give me some feedback on the clarity of
explanations such as this before then.	Thanks.

-Howard