One possibility is a defunct connection with the waste or sheath tank. Check the status box to see if the settings are correct. If is says that the sheath is empty or the waste is full then the wiring may have broken. You can buy a new piece or bypass the interlock with a paperclip. Just one possiblity but it happened to us about two weeks ago. Brian Newsom Director, Flow Cytometry Baylor College of Medicine Center for Cell and Gene Therapy -----Original Message----- From: Dhanwanthie Ramduth [mailto:Ramduthd1@nu.ac.za] Sent: Friday, May 11, 2001 9:35 AM To: cyto-inbox Subject: facs calibur dear flow cytometrists we have been experiencing a problem with our facs calibur recently. the cytometer cannot switch from the standby to ready mode. when attempting to acquire data, it always comes up withn a warning message " cytometer status not ready". we have folowed the trouble shooting guide, and and always ensure taht the machine is pressurised, and all connections are intact. what could be the cause of this problem? and how do we solve it? kind regards danni