Re: Paraformaldehyde and PI

From: Mark Kukuruga (kukuru@med.umich.edu)
Date: Wed Apr 18 2001 - 17:55:42 EST


Tomas,
Fixation will eliminate the barrier to PI entering the cell.  But fixed cells are dead,
so should stain anyway.
You need to add the PI before fixation.
Some may suggest that PI is retained (somewhat) following fixation.  I don't know that
I believe this . . . I'd have to try this and see.  If you are interested in detecting
cells that were dead prior to fixation, use EMA.
See: "Riedy MC, et.al.  Use of a Photolabeling Technique to Identify Nonviable Cells
in Fixed Homologous or Heterologous Cell Populations.  Cytometry 12:133-139, 1991."
MAK.

--
Mark A. KuKuruga, Managing Director
University of Michigan Flow Core
7416 CCGC 0946
(734) 647-3216, fax (734) 936-7376
kukuru@umich.edu


>>> "Tomas Corcoran" <mascor@gofree.indigo.ie> 04/17/01 02:44PM >>>
I wonder whether anyone has experience or opinions on this :

Does 0.5-2 %paraformaldehyde fixation of endothelial cells or tumour cells permeabilise
cells to PI or not.Different sources say yes and no , but we have found that after
removal of adherent cells from culture plates with 2mM EDTA and fixation that almost
all cells stain with PI . Does paraformaldehyde fixation preclude meaningful viability
assessment or not .

Many thanks.

mascor@gofree.indigo.ie



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:15 EST