Re: Viability PI vs. AAD - concentration and compensation

From: Howard Shapiro (hms@shapirolab.com)
Date: Thu Apr 05 2001 - 07:10:21 EST


Kerstin Buettner wrote-

>I am stuck with a very basic problem. I am trying to separate dead from
>viable cells. People in my lab have always done this with 10µgPI/ml plus
>RNase for 20min room temperature. I have tried two other protocols: PI
>2µg/ml 5min on ice and AAD 1µg/ml 30min on ice. Every protocol yields
>different counts of dead cells - from 16 +/-3% (AAD), 22+/-7% (PI) to
>27+/-5% (PI combined with RNase).
>Is this a probem of concentration - are DNA-Dyes titreable like an antibody?
>Is it a question of temperature or time?
>
>We do not understand this, because we always thought that a dead cell is
>dead, has permeabilised membranes and therfore takes up the dye - not
>depending on concentration. We always thought that higher concentration
>only gives you a better separation of peaks.

It is true that dead cells with permeabilized membranes should take up
otherwise impermeant nucleic acid dyes such as 7-AAD and PI.  Note that
propidium (PI) and ethidium (EB) should not be used interchangeably;
ethidium has only a single positive charge, and can get into intact cells,
especially at relatively high pH, but is pumped out.  Propidium has two
positive charges and generally does not get in.  TO-PRO-1 and TO-PRO-3 also
have two positive charges, and, in my limited experience (unpublished)
using PI (488 nm excitation) and TO-PRO-3 (633 nm excitation) together,
cells that take up one dye take up the other.

However, my colleagues and I have also observed (Novo et al, Antimicrob
Agents Chemother 2000, 44:827-834 [should be on PubMed]; Shapiro, Cytometry
2001, 43:223-226 and Shapiro, J Microbiol Methods 2000, 42:3-16 [this is in
a special issue on Microbial Analysis at the Single Cell Level, and it and
all other articles are available free at
http://www.elsevier.com/locate/jmicmeth]) that, under some conditions,
bacteria that maintain their membrane potentials, and which therefore have
intact membranes, take up PI or TO-PRO-3.

The "permeabilization" that permits PI uptake is usually conceived as a big
hole in the membrane, which should also allow equilibration of ion
concentrations across the membrane, reducing membrane potential to
zero.  This is clearly not the situation just described; the PI or TO-PRO-3
would have to get in via some more specific mechanism, e.g., a transport
protein, and I have heard that such transport proteins have been
described.  Suffice to say that the "membrane permeability" which is taken
as evidence of nonviability when dye exclusion tests are used may be
substantially more complex than was originally thought.

At a simpler level, there is the possibility that the different
experimental conditions you use could affect the development of membrane
permeability in dying (e.g., apoptotic) cells; while early apoptotic cells
exclude PI and can be identified only by other criteria (Annexin V binding,
mitochondrial membrane potential changes), these cells eventually end up
permeable to PI.  Along the way, you may see weak staining, so you have to
decide what is "positive".

I don't think enough has been published on this topic; if the numbers you
report hold up in a repeat experiment, you should put together a manuscript.

>How do I know which concentration is the right one for my AAD?

That would depend on whether you can resolve your initial problem or
not.  Under normal circumstances, you shouldn't need to titrate nucleic
acid dyes for dye exclusion tests.


>And there ist still another problem with PI: If I stay with PI - is it more
>accurate to compensate PI out of FL2 (which in fact means 99,9%
>compensation) or is it better to leave the PI signal where it is, since
>the  dead cells are gated out anyway before the PE stained cells in FL2 are
>analyzed?

I'd leave the signal as is, since it puts the dead (or permeabilized) cells
off scale and lets you gate them out.

-Howard



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:14 EST