Viability PI vs. AAD - concentration and compensation

From: Kerstin Büttner (k.buettner@uke.uni-hamburg.de)
Date: Wed Apr 04 2001 - 06:14:01 EST


Hello flowers,

I am stuck with a very basic problem. I am trying to separate dead from
viable cells. People in my lab have always done this with 10µgPI/ml plus
RNase for 20min room temperature. I have tried two other protocols: PI
2µg/ml 5min on ice and AAD 1µg/ml 30min on ice. Every protocol yields
different counts of dead cells - from 16 +/-3% (AAD), 22+/-7% (PI) to
27+/-5% (PI combined with RNase).
Is this a probem of concentration - are DNA-Dyes titreable like an antibody?
Is it a question of temperature or time?

We do not understand this, because we always thought that a dead cell is
dead, has permeabilised membranes and therfore takes up the dye - not
depending on concentration. We always thought that higher concentration
only gives you a better separation of peaks.

How do I know which concentration is the right one for my AAD?

And there ist still another problem with PI: If I stay with PI - is it more
accurate to compensate PI out of FL2 (which in fact means 99,9%
compensation) or is it better to leave the PI signal where it is, since the
dead cells are gated out anyway before the PE stained cells in FL2 are
analyzed?

Thanks very much for any comments on these issues!

Kerstin
Kerstin Buettner
Institute of Biophysics and Radiobiology
University of Hamburg
Martinistr. 52
20246 Hamburg
Phone: 0049-40-42803-2938
Fax: 0049-40-42803-5139
E-Mail: k.buettner@uke.uni-hamburg.de



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:14 EST