RE: APC-conjugated antibodies

From: Len Brown (lbrown@mmri.mater.org.au)
Date: Mon Jan 15 2001 - 18:09:43 EST


Dear Sathi

You cannot use APC on the Epics XL. APC requires a red laser source which
the XL does not have. Standard set-up of the XL allows use of FITC, PE,
PE-Texas Red (ECD)and Pe-Cy5 in that order. There are a number of companies
now selling Pe-Texas Red conjugated antibodies.

Len Brown
MMRI
Brisbane, Australia

-----Original Message-----
From: Sathi [mailto:sathi@repromedix.com]
Sent: Saturday, January 13, 2001 6:01 AM
To: cyto-inbox
Subject: APC-conjugated antibodies



I like to do a 4 color-analysis on Coulter-Epics/XL for lymphocytes.   For
one of the marker, I have an APC-conjugates antibody.  Could any one tell
me the emission/excitations spectrum for APC-fluorochrome and which channel
should I use for APC on Coulter Epics/XL cytometer.   I will be using FITC
and PE for other channels and the info would help me to choose the fourth
color (ECD or PC-5).

I would appreciate any helpful comments.

Thanks
Sathi
-------------------------------------------------------------
T. Sathiyaseelan, B.V.Sc., Ph.D.
Repromedix Corp.
sathi@repromedix.com