DNA staining

From: DARZYNKIEWICZ ZBIGNIEW (DARZYNK@nymc.edu)
Date: Wed Dec 20 2000 - 18:42:40 EST


Binding of a (ligand) fluorochrome to DNA essentially follows the law of
mass action. One can calculate approximate concentrations of the ligand and
relate to the cell number to find out the range when there may be a
significant decrease in free ligand concentration that may affect DNA
stainablility. There is 3x10E9 DNA base pairs per cell (diploid cell in G1).
Most intercalators reacting with free (naked) DNA at saturation bind every
second base pair. Thus, potentially, in a single cell there is 15x10E8
binding sites. However, because within the cell some  nucleosomal DNA is
inaccessible to the intercalators (Cytometry, 5:355-363,1984) only a
fraction of the potential binding sites (10-70%, depending on the ligand)
actually can bind the ligand. Thus, in a single cells there is maximally
between  1,5 to10 x 10E8 sites that bind these ligands. Assuming average MW
of most DNA fluorochromes to be about 300, one can estimate (from Avogadro
number) that at a concentration 100 uM (30 ug/ml) in one ml there are
6x10E16 molecules of the ligand. One million cells has between 1.5 to
10x10E14 binding sites. Thus, there is an nearly 100 times excess of the
ligand per binding site when one million cells are stained in 1 ml volume at
100 uM dye concentration. One would expect that under these conditions a
change in cell number from 1 to 2 million (which will change a concentration
of the free ligant by 1%) should not be reflected by greater than 1% change
in stainability of DNA. However, at lower dye concentration (e.g. below 20
uM) or when cell number is more drastically changed (e.g. from 1 to  5
million) the change in DNA stainability, clearly, is noticeable. Needless to
say, if cells have higher DNA content, (tetraploid, arrested in G2/M) the
equilibrium is shifted towards lesser concentration of free dye, which also
may lead to lowed DNA stainability.
Happy Holidays everyone,
http://www.maybeitzme.com/snowshow/snosho.htm
Zbigniew Darzynkiewicz
Brander Cancer Research Institute
New York Medical College
19 Bradhurst Ave.
Hawthorne, NY 10532
tel: 914-347-2801
fax: 914-347-2804
http://www.geocities.com/z_darzynkiewicz



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:41 EST