Re: best 4-color combinations

From: Tom Mc Closkey (thomasm@nshs.edu)
Date: Tue Nov 07 2000 - 12:21:33 EST


Avis Peters wrote:
>
> We are looking into performing 4-color immunophenotyping analysis
> (leukemia/lymphoma) that must be compatable on a Coulter Elite and BD Calibur.
> We want to use FITC,PE, and APC and are wondering which would be the best 3rd
> color-PE/Cy7, Percp/Cy5.5, APC/Cy7, ECD, or other fluorochomes?

	It will be difficult to find get a 4 color combo that will work on both
machines.  Thae Calibur is designed for FITC/PE/perCP/APC and a 4 color
combo which works on the Elite is FITC/PE/ECD/APC.  To use ECD on a
Calibur you will need to change filters and to use a perCP or PECY5 with
APC on the Elite you need to put the data into a program capable of
software compensation.	The Cy7 tandems show up very poorly on the Elite
with stock PMTs.

good luck,
Tom

*****************************************************************************