Re: Clumping cells

From: Tom Mc Closkey (thomasm@nshs.edu)
Date: Tue Oct 24 2000 - 11:14:09 EST


Jeff Barry wrote:
>
>> A colleague is trying to look for apoptosis using the sub G1, Propidium
> Iodide, method however, his cells stubbornly refuses to form single cells
> and continue to aggregate causing our FacsCalibur to block every time we put
> the samples on. Can anyone relieve our misery by suggesting any sample
> preparation tips that will prevent cell clumping.
>

	A variation on this assay utilizes hypotonic lysis to isolate
individual nuclei instead of analyzing whole cells.  It is described in:
J Immunol Meth, 139: 271-279, 1991.  This should eliminate cell/cell
binding.

Good luck,
Tom

*****************************************************************************