Re: Clumping cells

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Tue Oct 24 2000 - 06:58:19 EST


Hi Jeff,

There are a number of tips to prevent cell clumping. Firstly, if they
are adherent or even if they grow in suspension but in clumps, make sure
that they are in as good a single cell suspension that you can make
before fixation. Inadequate ethanol fixation is a primary cause of
clumpage - its a good idea to vortex while fixing (careful though if
looking for apoptotic cells as they could explode at this point!). Also
make sure that the pellet is as free as possible of residual PBS before
adding the ethanol.

Secondly you could try using either EDTA or a low level concetration of
DNase in your PI solution as it is DNA from blown-up cells that can
cause clumps as well.

Also it can be a good idea to filter the cells through a 60-70um mesh to
get rid of the larger lumps (although clearly this will lose cells).
Finally, if they are someone elses samples, make sure you watch them
as they prepare them!

Good luck,

Derek


On Mon, 23 Oct 2000, Jeff Barry wrote:
> A colleague is trying to look for apoptosis using the sub G1, Propidium
> Iodide, method however, his cells stubbornly refuses to form single cells
> and continue to aggregate causing our FacsCalibur to block every time we put
> the samples on. Can anyone relieve our misery by suggesting any sample
> preparation tips that will prevent cell clumping.

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk
London, UK			   mobile: 07790 604112

Web Page: http://www.icnet.uk/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:36 EST