Re: magnetic beads and/or sorting

From: Hans-Peter Steffens (Hans-PeterS@miltenyibiotec.de)
Date: Wed Aug 16 2000 - 09:12:50 EST


on 11.08.2000 10:21 Uhr, Carolyn Jefferiss at prscmj@bath.ac.uk wrote:

Dear Carolyn,
I can imagine that you have to digest the information on the fluorescence
labeling of MACS separated CD34 cells.

> 1) Thanks you so very much, everyone who had replied to my
> minimacs/sorting queries, which I have yet to digest properly

The answers seemed to be contradictory so I will try to shed some light onto
the mystery.
>>
>> Appenrently CD34
>> labelling after MACS selection does not diminish the fluoresence
>> intensity of FITC, PE or APC labelled CD34+ cells.

>>> 2. If you are checking the purity of your cells with an antibody recognizing
>>> the same epitope as the MACS beads, you will have problems as many or most
>>> of the epitopes will already be occupied with the antibodies attached to
>>> the beads. You need another antibody which does not compete with the first
>>> one.

Both considerations were right - depending on the antigen.
The MACS MicroBeads work under extremely strong magnetic fields, thus you
really need only a few MicroBeads per cell to get a cell retained in a
separation column. For e.g. CD4 on human T cells is present at several
thousand molecules per cell, only a small proportion of these molecules  are
occupied by MicroBeads. In such a case it is possible to use the same clone
as a fluorochrome conjugate to stain the cells for flow cytometry.

In the case of CD34, the antigen is expressed at a lower level and thus the
Beads cover already a significant amount of the available molecules. The
antibody used for detection should recognize a CD34 epitope different from
the epitope recognized by the antibody which is conjugated to our
MicroBeads. The antibody clone QBEND/10 used for our CD34 Progenitor Cell
Isolation Kit and CD34 MultiSort Kit binds to epitope II of the CD34
antigen. For the staining of MACS isolated CD34+ cells we recommend using an
antibody recognizing CD34 epitope III, e.g. clone HPCA-2 (Beckton Dickinson,
San Jose, CA) or MACS fluorochrome conjugated antibodies CD34-PE or
CD34-FITC (clone AC136). Special recommendations are indicated in the data
sheet when necessary.

I hope this helps

Hans-Peter

--
Dr. Hans-Peter Steffens
Technical Support
Miltenyi Biotec GmbH
Friedrich-Ebert-Str. 68
51429 Bergisch Gladbach
Tel: +49-2204-8306-264
Fax: +49-2204-85197
http://www.MiltenyiBiotec.com/