Robert Adair McGilp wrote: > I have an investigator who has been using a FACSort to analyse his FITC/ > PE/Cychrome labelled samples and gets good resolution of the populations > of interest (dual positives). He feels that we have been unable to > reproduce the results on our Vantage SE. We have moved the z axis so that > the nozzle is a good inch above the laser intercept and this gives a good > approximation of the results he has seen on the analyser. Be sure to insert the appropriate filter in front of the FL3 detector. Pe-Cy5 has its emission maximum at 675 nm. BD has a 675/40 bandpass filter and a 650 longpass. Both should give a good resolution. In the standard configuration you often find a 630/22 or 610/20 designed for propidium iodide in front of FL3, which will not give good intensities for PE-Cy5. Jens