Re: facscalibur biocontainment

From: Leonard, Thomas (tleonard@wistar.upenn.edu)
Date: Wed Mar 15 2000 - 15:45:19 EST


A general tenet among safety professionals is that potentially hazardous
materials should be controlled at the source, or as close to the source as
possible by engineering and/or administrative contols. Personal protective
equipment is generally considered to serve as a secondary protection
measure and should be introduced as a primary protection measure only when
engineering controls are insufficient or temporarily unavailable.
Accordingly, I would be leery of incorporating "BL3 gear" (I assume PAPR,
or HEPA respiratory protection, etc) alone; particularly when addressing an
agent clearly transmitted via aerosol such as mycobacterium.

Second, although the room ventilation system may have been designed to draw
air away from the BL3 cytometer operator, this probably isn't alone
sufficient in protecting the operator considering that eddy currents and
other quantitative ventilation issues must be addressed. To illustrate,
place some dry ice in a container of water on the cytometer bench, then
open and shut the lab door, or walk behind, or stand in various positions
behind the cytometer. The visible currents tell a tale.

The suggestion to position the sorter in a laminar flow hood is, in theory,
a good approach; however this may be impractical.

Along with standard BL3 precautions, if your sort sample collection drawer
is under negative pressure (not so much to affect the sort stream) and
enclosed during sorting, you're off to a good start. I would suspect that
this feature, coupled with the secondary containment of a laminar flow
hood, or a properly designed room ventilation system perhaps with the
requirement of personal protective equipment (i.e. respiratory protection),
should be more than sufficient to protect research personnel and satisfy
your biosafety officer.

Regards, Tom






>>>To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>


>just to let those who were interested know, i got a few responses on this
>question.
>
>two individuals suggested that gowning up in complete BL3 gear was
>sufficient, since the cytometer was not supposed to generate aerosols.
>another said that her BL3 cytometer was located downstream of the airflow,
>so that air moves away from the operator.
>
>i'd like to hear if from BD reps to see if any of their customers have
>taken any extra precautions.  please reply to the list as lots of people
>seem to be interested!!
>
>thanks once again for the responses!
>
>barbara
>
>Barbara A. Butcher, Ph.D.
>Senior Research Associate
>Department of Microbiology and Immunology
>C5-160 Veterinary Medical Center
>Tower Road
>Cornell College of Veterinary Medicine
>Ithaca, NY 14853
>607 253-3270
>
>
>

********************************
R. Thomas Leonard, M.S.,CSP,CBSP
Safety Officer
The Wistar Institute
3601 Spruce Street
Philadelphia, PA 19104
tleonard@wistar.upenn.edu
Ph:215-898-3712
Fx:215-898-3868
********************************



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:12 EST