Greetings to all, I initially posted a response privately to this question, but I would like to ask a general question of those much more knowledgeable than I. Until recently I had assumed that pressure, sheath fluid, and collection medium were the main causes of cell death after sorting. However, recently we had a problem with stray voltage on a FACStarPlus (shocks after just touching the cytometer, plates charging by themselves, etc.) that was coincident with lousy viability after sorting. In frustration, I performed the following: 1) I collected the waste stream (pressure, no light, no electrical field)--cells were 97% viable, 2) I collected the waste stream (pressure, laser engaged, plates on)--cells were 96% viable. 3) I sorted cells--10% viable! (All of these tasks were performed in less than a 5 minute period). We had zapped the cells. Is it possible that the electrical charge behaves somewhat like electroporation which is really hard on cells? The manufacturer was stumped, but a retooling of the instrument solved the problem. I have never seen a discussion on this aspect of sorting, but could this also contribute to loss of viability and function? If so, can we adjust this? Thanks, Kathy -------------------------------------------------------------- Kathy Schell Supervisor, UWCCC Flow Cytometry Facility 600 Highland Ave. K4/535 Madison, WI 53792 Voice: 608-263-0313 e-mail: kschell@facstaff.wisc.edu