Re: Calcium Flux help

From: Ray Hicks (rh208@cus.cam.ac.uk)
Date: Thu Mar 02 2000 - 10:53:54 EST


Hi Brian,

What stimulant are you using? Are you using a concentration that you know
should give an appreciable change in fluorescence (not necessarily the same
as a "physiologically stimulating" concentration)?

If the answers to these are yes, it might be worth looking at your sheath
buffer, Isoton II (and I think FACSflow) can suppress calcium responses to
a range of agonists at quite low concentrations, e.g. if they get a chance
to drip back into your sample while you're removing it from/replacing it on
your machine.

I found that putting the machine (Vantage or FACStar) in standby works
fine, since it pinches the tube and stops drip-back, but using PBS as a
sheath is easier and more reliable.

Are you using a bench-top analyser or a big sorter?  If the latter, it
could be that the response is over by the time the cells reach the laser -
try boosting after you replace the tube.

Ray

 At 6:38 pm +0000 25/2/00, Newsom, Brian S. wrote:
>We tried our first stab at calcium mobilation today and got dismal
>results.
>Following is the protocol we used:
>
>Preparation:
>1.	Make up fluo-3 and Fura Red at 10mg/ml in DMSO
>
>Procedure:
>1.	Wash cells in Ca++ Free media (HBSS)at RT.
>2.	Incubate cells in media containing 16uM fluo-3 + 40uM Fura Red for
>45 minutes at 37C.
>3.	Wash cells once in Ca++ Free media (HBSS)at RT.
>4. Resuspend in HBSS (Ca++ Free) at 2.0 X 10^6/mL
>5.	Run baseline analysis on Flow Cytometer collecting 30s-1min
>of data.
>Collect fluo-3 in FL1 and Fura Red in FL3. Collect both vs. Time.
>6.	Take sample off machine and add stimulus (leave machine in
>run-continue collecting listmode).
>7.	Put sample back on and collect through sample peak (about 5 min).
>
>FL1 and FL3 were collected in linear. No flux was seen with either
>dye. Any
>suggestions on what we did wrong? Do we need HEPES in the buffer? Do
>we need
>different dye concentrations? Any suggestions welcomed and appreciated.
>
>Brian Newsom
>Director, Flow Cytometry
>Center for Cell and Gene Therapy
>Baylor college of Medicine


			      Ray Hicks
________________________________________________________________________
|University of Cambridge	  |Tel		    01223 330149	|
|Department of Medicine		  |Fax		   01223 336846		|
|Level 5, Addenbrookes Hospital   |e-mail	  <rh208@cus.cam.ac.uk> |
|Hills Road Cambridge		  |Web	http://facsmac.med.cam.ac.uk	|
|CB2				  |ftp server  ftp://131.111.80.78	|
|UK				  |					|
|_________________________________|_____________________________________|



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:10 EST