Re: Streptavidin

From: Steve McClellan (xenoman@home.com)
Date: Fri Feb 25 2000 - 20:11:31 EST


If you can tolerate several washes, you can use a biotin on the Ab, then an
unlabeled streptavidin step (1ug/10x6 cells).  Wash and add labeled biotin
at same concentration.

Avidin binds four biotins, this sandwich allows you to get 3 labeled
molecules per Ab.  I have had great success with it, at the level of 900-1800
target molecules per cell.

Good luck,
Steve McClellan
BeckmanCoulter cytometry

----- Original Message -----
From: Jeff Haug <jhaug@im.wustl.edu>
To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
Sent: Thursday, February 24, 2000 6:40 PM
Subject: Streptav. Second makes signal brighter?


>
> I have always been told that a biotin conjugated primary antibody followed
> by a fluorescent conjugated streptavidin secondary antibody is a method
> that should make a low density signal more bright.
>
> Is this really true or just theoretical?
>
> I ask this question because when I have performed comparisons between
> direct and indirect staining methods I have never seen much difference.
>
> Thanks, Jeffrey
>
>



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:09 EST