RE: Kupffer cell Ab labeling

From: thomasm@nshs.edu
Date: Mon Feb 11 2002 - 13:32:03 EST


> -----Original Message-----
> From: Nathalie Tessier [mailto:tessien@ircm.qc.ca]
> Sent: Friday, February 08, 2002 8:52 AM
> To: Cytometry Mailing List
> Subject: [Fwd: Paul Jolicoeur]
>
> >  We are having problems labeling the liver Kupffer cells
> (macrophages)
> > with different antibodies by FACS.  The isotype controls
> label the cells
> > as intensely as the antibodies .  As you know, these cells
> are designed
> > to actively phagocytose.  Is there a trick to get specific labeling
> > easily?

Hi Nathalie,

I would suggest the following steps in your procedure:

1]  Try not to use whole Ab to label, better to use F(ab')2 fragments
2]  Use an Fc blocking step prior to Ab labeling
3]  Keep the cells on ice throughout to prevent Ab internalizaiton
4]  always run unlabeled vs. isotype, and then isotype vs test so that the
amount of nonspecific labeling is measured for each experiemnt
5]  Don't use protease as part of your isolation procedure as this can alter
the cell surface

good luck,
Tom

****************************************************************************
*