Re: 7-AAD/PE problem

From: DAVID M CODER (d_coder@MSN.com)
Date: Wed Jan 30 2002 - 17:19:39 EST


----- Original Message -----
From: "Corver, W.E. (PATH)" <W.E.Corver@lumc.nl>
To: cyto-inbox
Sent: Tuesday, January 29, 2002 9:06 AM
Subject: RE: 7-AAD/PE problem


>
> Dear Brian,
>
> [part deleted]
> After fixation you can use methanol, ethanol, lysolecithin or detergents.
> For nuclear staining, we obtained the best results with methanol.
> Second, we also observed reduction of FITC as well as PE fluorescence
using
> PI and TO-PRO-3 as DNA stain. (Cytometry 15:117-128, 1994 and Cytometry
> 28:329-336, 1997). The phenomenon could not be fully explained by
absorption
> or energy transfer. This was very limited as demonstrated by fluorometry
> experiments using dye mixtures at the lab of Hans Tanke (Sylvius Lab,
> Leiden). What is causing the effect is still unclear to us but maybe
> somebody else has a suggestion.
Energy transfer seems the best explanation. FITC and PI are a near optimal
pair; PE and PI less so, but still there is overlap. PI and TO-PRO-3 are a
very good pair: the emission of PI overlaps well the absorption of TO-PRO-3
(e.g., a HeNe laser at 633nm excites TP-3 very well), and both dyes
intercalate into DS-DNA. In fact, the only way to observe TO-PRO-3
fluorescence when only a 488nm line is available is by transfer from PI. All
such energy transfer (FITC, PE --> PI; PI-->TP3) is avoided with direct
excitation of TP-3 by red (633 to 647nm) excitation.

Best regards,
Dave
----------------
David M. Coder, Ph.D.
Consultant in Cytometry
email: d_coder@msn.com
tel./messages: 206-499-3446




> -----Original Message-----
> From: Newsom, Brian S. [mailto:BSNEWSOM@txccc.org]
> Sent: maandag 28 januari 2002 22:02
> To: cyto-inbox
> Subject: 7-AAD/PE problem
>
>
>
> Question to all of you DNA experts. We have someone who is tring to run an
> experiment staining a nuclear localized protien with PE and then staining
> with 7-AAD. The 7-AAD staining works well with a good CV. The antibody
> staining looks decent (although low percentage and fairly dim) when by
> itself, but when the 7-AAD is added the antibody staining totally goes
away.
> Is there an energy transfer or steric hinderance issue that may be going
on?
> Any help appreciated.
>
> Brian
>



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:59:22 EST