Andreas, I believe all the "green" GFP variants are UV-excitable. Is the GFP fluorescence showing up in the Indo-1 "blue (long)" detector (~525BP)? Perhaps you could try DsRed instead of GFP. Does your system properly measure calcium responses with other cells? I have successfully run indo-1 samples on a FACS Vantage, and indo-1 with PE, but I have not tried indo-1 with GFP. Good luck, Eric >Hi everybody, >we have a BD FACS-Vantage and have tried to measure >Ca-fluxes in GFP overexpressing cells. The cells used are >a low-adherent tumor cell line and were detached by >little mechanical force (tapping the flask). We used INDO >as the Ca-dependent fluorescent dye. >Unfortenately, we had bad luck. Even with ionomycin, we did >not get a Ca response. >Has somebody experience with this combination (GFP INDO) >and can help us? > >A second question: >Who knows a flow cytometry company working about blood >homeostasis (platelet activation ...) in Europe? > >Thanks in advance > >Andreas > > > >PD Dr. Andreas Simm >Universitaet Halle Wittenberg >Klinik fuer Herz- und Thoraxchirurgie >Ernst-Grube Str. 40 >D-06120 Halle > >Tel.: +49 (0) 345 557 2647 > 552 2878 >FAX: +49 (0) 345 557 2782 > 552 2890 Eric Van Buren, aa9080@wayne.edu Karmanos Cancer Institute and Immunology & Microbiology Wayne State University, Detroit, Michigan, USA