Kevyn Adler, One of our clients wanted us to conduct endotoxin-free sorts on a FACS Vantage. Our procedure entailed preparing the cytometer with the following run sequentially through both the sheath tank and the sample port: 1% ES 7X, 10% bleach, 70% ethanol, distilled water, and finally "blood bank" quality saline. We received no indication from the client that endotoxins were ever present in the sorted samples, although I do not know if the samples were tested for endotoxin. After cleaning your FACScan, I would recommend installing a new sheath filter before the final saline wash. We use saline from Bio Test Medical, available from Fisher Scientific, catalog number MSC1000B. Good luck, Eric >All, > >I am looking for a good method to remove or destroy endotoxins from my >FACScan. I am not sure if 10% Bleach will suffice. Please advise me on >what other people use. > >Thanks you, > >Kevyn Adler > >If you have any questions please do not hesitate to contact me. > > >Sincerely, > >Kevyn Adler >Technical Marketing Scientist >Miltenyi Biotec, Inc. >12740 Earhart Ave. >Auburn, CA 95602 >Tel: (800) 367-6227 >Tel: (530) 888-8871 >Fax:(877) 399-5701 >*********************************************************************** > >Please visit our new web page and discussion forum at: > >www.miltenyibiotec.com > >*********************************************************************** Eric Van Buren, aa9080@wayne.edu Karmanos Cancer Institute and Immunology & Microbiology Wayne State University, Detroit, Michigan, USA