From: William Telford (TelfordW@mail.nih.gov)
Date: Sun Nov 18 2001 - 20:12:11 EST


Hi Lora...

Maybe your users originally used ECD / PE-Texas Red on an older FACScan or
FACSCalibur - our rather "ancient" FACScan (1987) has a 650 LP in front of
the FL3 detector, and PE-Texas Red conjugates work acceptably on it for
dense antigens, losing less than a log decade signal on average compared to
PE-Cy5. I think our FACSCalibur (also somewhat older, 1994) also has a 650
LP on the FL3 detector.  If the newer FACSCaliburs now have a 675 LP in
front of FL3, PE-Texas Red would be far less useful for this particular
instrument.

Anyway, you are absolutely right - why use PE-Texas Red on a Calibur and
risk losing signal intensity when so many PE-Cy5 conjugates are
available.  Your users should save ECD for the Vantage or for
Beckman-Coulter instruments, which are designed for it

Take care,

Bill

At 05:51 PM 11/14/2001 -0800, Barsky, Lora wrote:


>Hi everyone,
>
>In our quest to run more colors on the sorter, we have been using Coulter's
>ECD (PE-TxRed) conjugates on our Vantage SE.  The vantage is equipped with
>Enterprise and HeNe lasers.  We do not have any means to cross-beam
>compensate.  So, plugging the Bandpass for RED613 (610bp20) into FL3  has
>made it possible to sort four colors on our system along with the
>traditional, FITC, PE and APC.  We have also chosen ECD because PerCP
>quenches about 1.5logs on our Vantage and PE-Cy5 crosses too much into APC.
>
>BUT...  I have users here who want it to work on the Calibur, because they
>don't want to buy or can't buy PerCP conjugates or PE-Cy5 conjugates and
>because they want to directly compare emissions from both machines. But that
>is another problem of mine - I digress... So, since we cannot change out the
>optics for FL3 on the Calibur, they are collecting only the far red tail of
>the emission with the standard 675LPfilter; they are completely missing the
>emission peak.  I've told them that the conjugate wasn't designed to work on
>a Calibur, but the user "saw signals" and now I'm dealing with their "it
>works fine for me" syndrome, when I ask them to refrain from promoting this
>antibody combination to other users who are beginning 4 color work.  Could
>anyone provide me references that would help get my point across?
>
>If there is someone out there who does this combination successfully on
>their Calibur and really trusts their data, could you share with me the
>typical instrument settings values you're coming up with? I understand that
>there are instrument to instrument differences but would still like to
>know...
>
>Thanks,
>Lora W. Barsky, Research Specialist
>CHILDRENS HOSPITAL LOS ANGELES
>Research Immunology/Bone Marrow Transplant
>Flow Cytometry Core
>4650 Sunset Blvd. Mail Stop #62
>Los Angeles, CA  90027
>(323) 669-5935