This is not a control too far! A substantial fraction of the intracellular cytokine staining that I see appears to be artifact. Blocked controls such as you describe are one of the best ways to verify your data and avoid such artifacts. Intracellular staining with a new mAb or new system should be validated 2-3 times using either an unlabelled mAb or recombinant cytokine blocked control. Isotype controls by themselves are not adequate and should only be used once you have performed more rigorous blocking controls. Different mAbs posses different non-specific binding characteristics, thus some isotype matched controls may poorly match the non-specific bonding of you anti-cytokine mAb. Fixed permeabilized cells demonstrate far greater non-specific binding than fresh cells and thus require more stringent controls. I have typically used 100 mcg/ml of unlabeled mAb (staining in 50 ul) to block and then add the labeled anti-cytokine mAb to the blocked cells WITHOUT washing out the unlabeled mAb. The correct positive control for this is to block with 100 mcg of isotype control, then stain with the labeled anti-cytokine mAb. The most rigorous control is to stain with 2 different mAbs that recognize 2 different mAbs on the intracellular Ag, demonstrate they recognize the same population and then block each separately (Prussin, J Immunol Methods 188: 117-128) A more detailed discussion of these issues is presented in Unit 6.24 of current Protocols in Immunology. Calman > _______________________ > Calman Prussin > Laboratory of Allergic Diseases > NIAID/ National Institutes of Health > Building 10, Room 11C205, MSC-1881 > Bethesda, MD 20892-1881 > (301) 496-1306 (Voice) > (301) 496-1306 (Fax) > calman@nih.gov > > ---------- > From: Finney, Simon > Sent: Wednesday, September 12, 2001 5:52 > To: Cytometry Mailing List > Subject: More about cytoplasmic staining. > > > Dear all, > > There has been a bit of discussion about intracellular staining recently > with primary conjugated antibodies. > > I was trying to prove specificity of my intracellular binding by blocking > with unconjugated antibody first. My conjugated antibody still appears to > bind but I am unclear as to appropriate doses to block with and whether > the > conjugated one can swap places with unconjugated one during the second > incubation. > > Is this a control too far ? > > Isotype controls (admittedly from a different company) show no binding. > > Thanks > > Dr Simon Finney > Unit of Critical Care > NHLI at Imperial College > London UK > >
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:53 EST