Re: 3 color fluorescense on a Scope?

From: Ronald Rabin (rr84g@nih.gov)
Date: Sun Aug 12 2001 - 10:34:52 EST


Jim, you cannot use PE on a fluorescent scope.  I know this is true for
confocal, and am pretty sure it is true for any fluorescent microscopy.  It
bleaches too quickly.  Three colors you can use are FITC, Cy3 and Cy5.  Pay
attention to the filter cubes, because there is spectral overlap between Cy3
and FITC on one end, and Cy5 on the other.  Also, you have too have a camera
and imaging system to "see" Cy5, as it is barely visible with the naked eye.

ron


on 8/10/01 12:33 PM, jim phillips at jphillip@med.miami.edu wrote:

>
> Hello to everyone,
> This is a little off the subject of flow... but I have an investigator
> that want to do 3 color cell surface fluorescense on a microscope.  The
> question is .. what three fluorochomes can she use..  The microscope is
> just a normal fluorescense scope.  I know FITC and PE but is there a
> third?? Can you use Cychrome???   She wants to view all three at the
> same time..  any help would be appreciated.. Thanks...
> Jim Phillips
> University of Miami School of Medicine.
>
> ps.... I just finished my music CD .. entitled ... "222.... 2 old?  2
> fat? 2 ugly?"  You can listen to it on mp3..
>
> www.mp3.com/jim_phillips
>

Ronald L. Rabin, M.D.
Senior Staff Fellow
Laboratory of Immunobiochemistry
DBPAP/OVRR
Center for Biologics Evaluation and Research
U.S. Food and Drug Administration
29 Lincoln Drive (MSC-4555)
Building 29, Room 129
Bethesda, MD   20892-4555

phone:  301.496.8806
fax:    301.402.5177
email:  rr84g@nih.gov



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:50 EST