Hi All I have been unable to detect a PharRed signal from samples I have been given, and don't know if the problem is instrument or labeling error. MoFlo set up is as follows: Hammatsu 957-08 PMT all around Omega Optical Filter Set XF501 (consists of 710 DRLP, 670DF40, and 740EFLP) Melles Griot 35mW HeNe Red laser alignment seems fine: "Molecular Probes AlignPlus beads for 633" are almost off scale if no ND filter, at 399V, linear gain 1. BD's APC like Calibrites show excellent separation + / - ( >2 logs) 580V. Whether I measure APC in FL5 (remove 710LP dichroic, use 670/40BP) or FL6 (710LP plus 670BP), I obtain good APC signal. But no PharRed with the 710LP dichroic / 740EFLP in FL5, and no PharRed signal even if I remove the 710LP dichroic. BD Pharmingen Technical data sheet recommends a Hammatsu R3896 PMT for improved red sensitivity. Is the Hammatsu 957-08 deficient in the >740 range? If not, what's up? Thanks, Claude