Contrary to a variety of posts to this list, while paraformaldehyde does indeed kill cells, it does not necessarily permeabilize them enough to allow PI into all of the cells. For rat bone marrow cells, less than 25% of the cells take up PI following 0.5% paraformaldehyde fixation (as evaluated under a fluorescent microscope), even though 100% of the cells are indeed dead. Additional permeabilization is needed to make all the cells pick up the dye. There may be some variability between cell lines, etc. It is ALWAYS best to actually look at the cells when a new method is tried. Carol W. Johnson, DVM PhD Pharmacia Kalamazoo, MI Maryalice Stetler-Stevenson <stetler@box-s.nih.gov> on 04/18/2001 03:31:37 PM To: cyto-inbox cc: (bcc: Carol W Johnson/USKZO/PNU) Subject: Re: Paraformaldehyde and PI PI is only excluded from live cells that are able to maintain their cell membranes. Paraformaldehyde does kill cells.Therefore, fixing live cells kills them. Then PI accurately demonstrates they are dead. Who ever told you otherwise knows absolutely nothing about PI exclusion from cells. Maryalice >I wonder whether anyone has experience or opinions on this : > >Does 0.5-2 %paraformaldehyde fixation of endothelial cells or tumour >cells permeabilise cells to PI or not.Different sources say yes and >no , but we have found that after removal of adherent cells from >culture plates with 2mM EDTA and fixation that almost all cells >stain with PI . Does paraformaldehyde fixation preclude meaningful >viability assessment or not . > >Many thanks. > ><mailto:mascor@gofree.indigo.ie>mascor@gofree.indigo.ie > Maryalice Stetler-Stevenson Director Flow Cytometry Unit Laboratory of Pathology, NCI, NIH "Learn the rules so you know how to break them properly." The Dalai Lama PI is only excluded from live cells that are able to maintain their cell membranes. Paraformaldehyde does kill cells.Therefore, fixing live cells kills them. Then PI accurately demonstrates they are dead. Who ever told you otherwise knows absolutely nothing about PI exclusion from cells. Maryalice I wonder whether anyone has experience or opinions on this : Does 0.5-2 %paraformaldehyde fixation of endothelial cells or tumour cells permeabilise cells to PI or not.Different sources say yes and no , but we have found that after removal of adherent cells from culture plates with 2mM EDTA and fixation that almost all cells stain with PI . Does paraformaldehyde fixation preclude meaningful viability assessment or not . Many thanks. <mailto:mascor@gofree.indigo.ie>mascor@gofree.indigo.ie Maryalice Stetler-Stevenson Director Flow Cytometry Unit Laboratory of Pathology, NCI, NIH "Learn the rules so you know how to break them properly." The Dalai Lama
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:34 EST