Please note that this approach (blocking with BSA-Cy5) will only work if you are not using a second laser with APC as a stain. If you use APC the compensation would become impossible (i.e. the background staining of the bound BSA-Cy5 would wreak havoc on the APC channel).. Alan Stall BD-PharMingen =================== Gina_Graziani-Bowering@hc-sc.gc.ca on 04/17/2001 10:54:36 AM To: cyto-inbox cc: (bcc: Alan Stall/SDCA/BDX) Subject: Re: PE-Cy5 Backgrounds Hi Doug, To avoid the problem of PE-Cy5 conjugated antibodies binding to Cy5 receptors on the surface of monocytes, we used to preincubate the cells with BSA-Cy5 reagent prepared by conjugating BSA (Sigma) to Fluorolink Cy5 Reactive Dye (Amersham Life Science, Inc., Pittsburgh, Pa.) as per the manufacturer's instructions. This process is analagous to blocking cells with gamimmune prior to staining. Hope this helps, Gina _______________________ Gina Graziani-Bowering, PhD Post-Doctoral Fellow Health Canada