Re: DAPI as a viability exclusion dye

From: Stefan Andreatta (stefan.andreatta@uibk.ac.at)
Date: Thu Mar 15 2001 - 12:55:39 EST


Hi,

This is an interesting question. I just interwiewed a colleague, who is
a specialist on photobiology. In comparison to UV-irradiation found in
nature, the does-rate in a cytometer is of course enormous while the
actual dose is minimal due to the short exposure. For comparison: the
UV-radiation measured at 3000m a.s.l may be approximately 600
milliWatt/square-meter. While the radiation in a 50micron focus of a
50mW laser is about 25 Megawatt/square-meter. Thus, mathematically, a
500 nanosecond flight through the laser would be equivalent to a 10
second sunbath in the mountains.

The DNA doesn't absorb at 350nm, so UV light at this wavelength
shouldn't cause direct mutation. On the other hand it can lead to the
production of radicals within the cell, which happens within
picoseconds.

However, to estimate from this basis the actual damage to a cell in the
cytometer is just speculation. Unfortunately, I don't know of any
investigations done specifically for this situation.

Best Regards

Stefan

-------------------------------------------------
Stefan Andreatta
Institute of Zoology and Limnology
University of Innsbruck, Austria
phone: ++43-512-507-6122; fax (..)-2930


Mark Kukuruga wrote:
>
> Hi all . . .
> Considering the use of DAPI as an exclusion dye, I'm wondering about the effects of
> (albeit brief) UV irradiation on the cells that we sort "viably."  Should we worry,
> or is it negligible?
> MAK.
>
> --