Re: Hi FLOWers; Temperature of incubation with Abs to cell surface antigens

From: Simon Hunt (simon.hunt@pathology.oxford.ac.uk)
Date: Wed Mar 14 2001 - 11:59:30 EST


Hi,

Temperature of incubation with Abs to cell surface antigens

Simon Monard gives the anti-capping reason and it is an excellent one.
Maybe those who prefer ambient temperature (whatever that may be in various
parts of the world) get away with it because some cell surface molecules are
very much less likely to cap than others.  Surface Ig on B cells caps in the
twinkling of an eye, as the elderly among us will recall from the early-70s
controversy about whether it was there at all.  This argument was resolved
by
incubating lymphocyte suspensions at 4 degrees during staining with anti-Ig
and keeping them cold.

There is an additional fundamental reason, which concerns Ab affinity
(strictly, avidity since practically everyone uses divalent antibody).
Where the values have been published, the off-rate constants for defined
antibodies such as monoclonals bound to cell surface molecules are about
ten-fold less at 4 degrees than at 25 degrees.  (There aren't many such
determinations, so I hope I'm not over-generalising).  For an example that
I'm familiar with, the half-time for dissociation of Fab of MRC OX7
(anti-Thy1.1;
sorry - anti-CD90) increased from 700 seconds at ambient (actually 18
degrees -
this is England!) to 7000 seconds at 4 degrees.  (Fab)'2 of W3/25 (anti-rat
CD4) is another example.  The message is to keep the cells cold (4 degrees)
not only during incubation but also during washing afterwards right up to
the addition of fixative.  People get away with ambient because (1) there
may be a compensating acceleration of on-rates at the higher temperature (2)
they think they don't mind losing a bit of signal (what you never see you
never miss) (3) some monoclonals may have sufficient oomph to keep hanging
in there even in heat waves.

If you want to read more on this, go to the chapter on "Kinetics of antibody
reactions and the analysis of cell surface antigens" by Don Mason and the
late Alan Williams.  It's Chap 38 in volume 1 of Handbook of Experimental
Immunology, ed Weir DM et al, Blackwells Scientific Publications, 4th
edition (1986) ISBN 0-632-00975-6 (the off-rate data I have quoted are in a
Table there).  When you read this, you will also come to challenge the
almost universal advice from commercial staining protocols about how much Ab
to add during cell labelling reactions (e.g. "one microlitre per 5 million
cells" or somesuch instruction).  What matters is the absolute concentration
of antibody molecules (in relation to the antibody affinity and the proposed
staining time), not the ratio of antibody to the number of cells in the
incubation mix.  In saturating antibody conditions, only a tiny amount of
the staining antibody latches on to the cell surface during an incubation
(much much less than 5% - the rest goes down the sink).  So upping the cell
number, within reason, makes very little difference.  It *does* matter how
much reaction volume you dilute your staining antibody into.  But all this
is a different point.  As usual, when you're setting up the protocol for
your favourite antibody, invest a little time and do the experiment on the
temp; Ab concn; staining time; cell number!

Don Mason has kindly seen this email before it's despatched to the
four corners by the e-pigeons, and is content with what it says.
He adds: "Ab on-rates can also differ widely [from one Ab to another].
W3/25 is x14 faster than W3/13 [anti-rat CD43], at least at 4 degrees C.
Consequently one may have to use some MAbs
at higher concentration than others, just to compensate for the low on-
rate. In fact we did not appreciate the slow on-rate for W3/13 when we
started the study that you mention. The finding that it was so slow
resolved the apparent contradiction that the site number for W3/13 on
lymphocytes was higher than that for W3/25 by absorption assays but lower
by binding assays....".  He reinforces the advice to do a few labelling
tests at first.

Simon Hunt
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr Simon V. Hunt, Lecturer in Immunology,  University of Oxford
Mail to: Dunn School of Pathology, South Parks Road, OXFORD OX1 3RE,  U.K.
Phone:+44/0 1865 275575 (office/voice mail) 275574 (lab). Fax: 275515
(general office) 285744 (lab personal)
http://dunn1.path.ox.ac.uk/~svhunt/labhomepage/SVHunt_WebHomeFeb2001v1.htm
Fellow in Immunology and Tutor in Medicine, Keble College
Mail to: Keble College, OXFORD OX1 3PG, U.K.
Phone:+44/0 1865 272749 (voicemail). Fax: 272705
http://senior.keble.ox.ac.uk/fellows/


----- Original Message -----
From: "Simon Monard" <smonard@trudeauinstitute.org>
To: cyto-inbox
Sent: Monday, March 12, 2001 9:18 PM
Subject: Re: Hi FLOWers,


>
> Hi
>
> People always used to incubate on ice for 30-60 minutes to try to prevent
capping, also
> to create a convenient time to have lunch or extended coffee and smoking
breaks. Nowdays
> you can add a tad of Sodium azide to prevent capping.
>
> Simon
>
> Simon Monard
> FACS Lab Manager
> Trudeau Institute
> Saranac Lake
> NY12983
>
> Ph 518 891 3080 X352
>
>
> >>> "Heather Medbury" <Heather_Medbury@wsahs.nsw.gov.au> - 3/11/2001
5:50PM >>>
>
> Hi FLOWers,
>
> About 5 years ago when I was analysing lymphocytes, we used to have the
cell preparations
> at 4 C for labelling with the antibody.
> The work I do now on whole blood, the cells are labelled at room
temperature.
>
> Why do people label at 4 C when it works perfectly well at room temp?
>
> Heather........