Re: High speed sorting onto a glass slide

From: Larry Arnold (lwarma@med.unc.edu)
Date: Fri Jan 26 2001 - 10:15:31 EST


Howard

I typical run this tip at 10-12 PSI.  DDF is 14-16K.  I usually limit the
sort speed to 33% of DDF so would try not to exceed 5K/s (unless of course
I'm wanting to go very fast in an enrichment mode - have done 10K/s on the
150 tip).  The calculated cell velocity would be around 10-11m/s.

Larry

At 08:53 PM 1/25/2001 -0500, you wrote:
>Larry-
>
>>Works fine.  Cells will be a bit dilute.    I have sorted with my MoFlo
>>onto a slide put on a cover slip and gone straight to my LSC.  Cells are
>>fine.  Usually put about 1000 cells down for a 22mm2 cover slip suing the
>>150um tip so could do more with a 70 or 100um tip.
>
>I guess I never thought about using a 150 um tip on a high speed
>sorter.  How many cells a second can you analyze with that big a stream,
>and what is the flow velocity?
>
>Thanks,
>
>-Howard
>