Re: Calcium flux experiments with standard BD Facscan/Facscalibur (instrument filters)?

From: Ray Hicks (rh208@cam.ac.uk)
Date: Fri Dec 29 2000 - 12:16:18 EST


Dear Ursula,

The laser and the filters in the FACScan/Calibur will let you down I'm
afraid - Indo-1 requires UV excitation to generate its violet and blue-green
signals for ratioing.  The usual strategy for 488nm work is to load the
cells with Fluo-3 (or similar), and ratio the increase of green fluorescence
against either the fixed red (isosbestic point) fluorescence of SNARF-1 or
the decreasing red fluorescence of Fura-Red.

It's best to switch time on as a parameter, although you can plot your ratio
against a "calculated" time parameter based on the order in which the events
are saved (naturally, this presumes an even and un-interrupted flow rate).

At this point I'd recommend my program FCSPress for your kinetic analyses,
it's very straightforward to use and it allows you to plot a variety of
statistics against time including mean, median and percent responding (it
even inserts time points automatically if you didn't acquire with a time
parameter) - and you'll find that you can transit to it quite rapidly.  You
can see an example kinetic plot (and other plots of a calcium experiment at
http://www.fcspress.com/html/FCSPressFeatures/graphics.html). You can
download a copy from http://www.fcspress.com which will run for thirty days.

Ray

Ray Hicks  134 High Street, Harston, Cambridge. CB2 5QD. UK
e-mail     RayH@FCSPress.com
Web        http://www.FCSPress.com
Tel         +44 797 453 8647
Fax        +44 870 740 8595----- Original Message -----
From: "Ursula Esser" <uesser@UCDAVIS.EDU>
To: cyto-inbox
Sent: Wednesday, December 27, 2000 8:25 PM
Subject: Calcium flux experiments with standard BD Facscan/Facscalibur
(instrument filters)?


|
| Dear Flowers,
|
| It's the holidays, and I have a hard time tracking down if I can do
calcium
| flux experiments (using Indo-1 AM or alike) by flow using the standard
| equipped BD Facscan or Facscalibur instruments (in particular: filters
ok?).
|
| As I have only performed steady state experiments, I am also not sure if I
| have to change the acquisition mode in order to allow analysis using time
| as a parameter (either in CellQuest or FlowJo (to which we are slowly
| transitioning)).
|
| Any suggestions, hints, info, recommendations for data analysis?
|
| Thank you all in advance for your help.
|
| Ursula Esser
|
| -- End --
|