Re: autofluorescence & isotype control fluorescence

From: Mario Roederer (Roederer@drmr.com)
Date: Fri Oct 20 2000 - 13:39:00 EST


Karim:

>Suppose you got highly autofluorescent cells and the isotype control is
>in the 2nd-3rd decade. The same isotype control in low autofluorescent
>cells is under the 1st decade.
>Now suppose the signal for marker X on these high autofluorescent cells
>is also in 2nd-3rd decade, and exactly overlaps the isotype control.
>
>What's the correct interpretation of these results?
>1) there's no difference between isotype control and marker X, so the
>cells are negative for marker X
>or
>2) marker X is in the 2nd-3rd decade, so it's definitely positive
>regardless of isotype control background

(1) is correct.  Although you really ought to have a completely
"unstained" control to make sure what autofluorescence is, that your
binding is not being caused by other factors (like Fc binding).
Nonetheless, that wouldn't change the interpretation that there is no
difference and the cells are negative, it would only aid in
understanding what's going on.

>Another way to state the question: are fluorochrome fluorescence (be it
>Ag-specific or aspecific binding) and autofluorescence additive signals
>for the same cell?

Yes, they are.

You might want to consider using Autofluorescence  Compensation to
reduce the contribution of autofluorescence in your channel of
interest.  To do this, you must have another fluroescence channel
that is "unused", and in which AF correlates well with AF in your
measurement channel.  Then you can use software or hardware to
compensate out some of the autofluorescence.  Since this compensation
will not reduce specific signal from the fluorochrome, you achieve
much greater sensitivity, and can sometimes detect even low level
signals against a high AF background.

For more information, see papers in 1986 or 1987 in Cytometry on AF
correction by spillover correction.  One is Roederer & Murphy, the
other is Alberti, Parks, & Herzenberg.

mr