Re: Calibrating for internal pH

From: Elizabeth R Simons (esimons@bu.edu)
Date: Thu Aug 24 2000 - 11:45:35 EST


We haven't ever done pH in microorganisms but have done so in human
blood cells - platelets, granulocytes, now doing it in monocytes, and in
phagocytic compartments (all are published). In all cases we used BCECF
and calibrated using nigericin. However, that has to be done in a buffer
system whose K+ concentration mimics that of the interior of the cell,
because nigericin collapses the ratio of H+in/H+out to equal that of
K+in/K+out, which therefore needs to equal one. That seems to be the
thing people who've contacted me in the past don't take into
consideration. If one takes that precaution, we have had very, very
reproducible calibration curves.
Elizabeth R. Simons

joerg ueckert wrote:
>
> Folks,
>
> We're holding us busy by applying CFDASE for ratiometric determination of
> internal pH in bacteria and yeast. Apart from strain- and pre-history -
> dependent difficulties in dye uptake the generation of a 'useful' calibration
> curve is the biggest challenge. The well-known combination
> valinomycin/nigericin results in a high degree of variability and curve shifts
> upon repetition. Also, apart from these, other agents (e.g. alcohols) or
> methods (e.g. electroporation) to facilitate equilibration of internal and
> external pH give stunning differences compared to mentioned 'standard' method.
> All remarks, suggestions and experiences with agents which provide reliable
> calibration curves for pHi in microorganisms are welcome. Providing a
> sufficient number of responses, I will post a summary on the list. Thanks in
> advance.
>
> - Joerg
>
> ------------------------------------------------
> Joerg Ueckert
> Unilever Research
> Microbiology
> Vlaardingen, NL
>
> Email   joerg.ueckert@unilever.com
> ------------------------------------------------

--
Elizabeth R. Simons, Ph.D.
Professor of Biochemistry
Boston University School of Medicine
715 Albany Street, K 602
Boston, MA 02118
phone (617) 638-4332
FAX   (617) 638-5339
email: esimons@bu.edu



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:56:02 EST