Date: Thu Jul 27 2000 - 17:48:12 EST


I've seen this before.  Tricolor/cychrome/PE-CY5, whatever you want to call it
is a tandem conjugate.  The idea is that the 488 laser excites the PE, it
fluoresces orange-red and this is in turn absorbed by the CY5 portion of the
tandem molecule and then it flouresces further red.  This energy transfer is not
100% efficient so you can get significant PE fluorescence showing up in FL2.
I'm gathering from your question that you don't have a problem compensating
FL2-FL3.  Compensation is always more difficult when you have two stains that
are vastly different in relative intensity.  The more disparity, the more
compensation needed.  Your FL4-%FL3 problem is exacerbated by the fact that it
isn't JUST a compensation problem.  Compensation is used to remove the
fluorescence contribution due to spectral overlap among fluors.  For instance,
you have to compensate your "green"(FITC/FL1) dye from the "orange"
(PE/FL2)channel because it has an orange "tail". We express this as FL2-30%FL1,
for example.  What we're saying is that an amount of light equal to 30% of the
light detected in FL1 is showing up in FL2 and we need to ignore that much of
the contribution to signal in that detector.  The problem you're seeing isn't
due to to spectral overlap.  You see, the CY5 portion of your tandem dye,
tricolor, is excited very well by your little red diode laser.  It's "real"
fluorescence in FL4.  (Imagine trying to compensate PE fluorescence out of FL2.
It doesn't work so well.)    So, the reason your FL4(APC?) signal is being
ablated is because it is lower than the siganl you are getting from tricolor.
If you compensate the tricolor signal in FL4, you are knocking your dim FL4
signal off scale. You are also correct that there is a time delay calculation
that has to be performed by the cytometer to ensure that signal collected when
cells strike the first laser is correlated with signal from excitation by the
second.   You should run FACSComp and do the time delay calibration every time
you are using the diode laser.  If you're not using FL4, turn it off.

And that's all I have to say about that.

David McFarland
SmithKline Beecham Pharmaceuticals
---------------------- Forwarded by David C McFarland/DEV/PHRD/SB_PLC on
07/27/2000 18:14 ---------------------------


David_Topham@urmc.rochester.edu on 25-Jul-2000 11:42



To: cyto-inbox
cc:    (bcc: David C McFarland/DEV/PHRD/SB_PLC)
Subject:  FL3/FL4 compensation on a Calibur





Dear flow cytometrists:

We have recently become aware of an interesting artifact regarding 4-color
samples on our Calibur.  The problem occurs when analyzing samples stained
with a bright FL3 fluorochrome, say CD8-tricolor, and a weak FL4 stain, say
CD25-biotin/SA-APC.  When compensation is set properly with the single color
reagents, the weakly staining CD25+/CD8+ population appears to be CD25
negative.  The brighter the stain in FL3, the more negative the FL4+ cells
appear.   So CD25 appears OK on the CD8-negative population.   We know this
is not correct because when another color is used for the CD25, say CD25-PE,
it appears just fine.  I believe there is a time delay for the FL4 detector,
and I wondered whether this factors into the compensation issue.   I have
also noticed that it is nearly impossible to overcompensate FL3-FL4, i.e.
make the events squish against the axis.  We are still experimenting with
the issue to try to resolve it with different settings, but I was hoping
someone could explain it to me a little more clearly.

David J. Topham, Ph.D.
Center for Vaccine Biology & Immunology
Aab Institute for Biomedical Sciences
University of Rochester Medical Center
601 Elmwood Avenue, Box 609
Rochester, NY 14642-8609
Tel. 716 273-1403
FAX 716 756-5614
E-mail: David_Topham@urmc.rochester.edu



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:58 EST