Tami, Whether or not you run FACSComp every day is dependent on your applications. If you are doing clinical or GLP type of work, you must run it every time you run the machine. However, contrary to BD's insert with the Calibrites, you can make the beads up, store them in the dark in the refrigerator, and use them at least two weeks if the bead lots are less than 6 months old. I have stretched the same beads up to a month, although that is not something I recommend if you want your instrument to pass every time. In my hands I have found the most labile of the beads to be the PerCP beads. I have never had any of the other beads fail, only the BD proprietary PerCP beads. (BD, are you listening?) Randy Fischer TherImmune Research Corporation 9700 Great Seneca Hwy Rockville, MD 20850 (240) 453-6256 RFischer@therimmune.com > ---------- > From: Tami Rosario > Sent: Wednesday, March 29, 2000 10:55 AM > To: Cytometry Mailing List > Subject: FACSComp QC > > > I have a question for those of you who run FACSComp on the FACSCaliber > for QC. Do you run this each time you start up the cytometer or only > occasionally. I have been told that I should run FACSComp each day > that > I run samples, but the beads are fairly expensive. Any alternatives? > > Thanks, > Tami > >