I discuss the problem with 3-color compensation on one of my pages,<http://www.drmr.com/compensation/3color.html>. (part of the "detailed discussion" on compensation). The fact is that one can never guarantee correct 3 color compensation using only pair-wise compensators (i.e.., FL1<->FL2 and FL2<->FL3) because they don't necessarily correct for FL1<->FL3 spillovers. Historically, this hasn't been too much of a problem because of the coincidental spectral properties of fluorescein and PE. There is no good solution to your problem other than to use software compensation (e.g., FlowJo on Macintosh or WinList on PC). If you use this approach, make sure that no pair of colors is overcompensated ("Squooshed"), because software can never correct over-compensation--it can only correct incomplete undercompensation. And, no, having less sensitive instrumentation will not fix this problem. By the way, I want to remind everyone that compensation settings DO NOT DEPEND on the brightness of a reagent! Therefore, you should always use the brightest reagent you have in any given channel to set compensation! mr