>>Colleagues,
>>
>>Recently we purchased a new FACSCalibur which runs Cellquest on a 400MHz G3.
>>I noticed that during initial acquisition of a sample there is a 'spurt' of
>>activity that
>>produces an image of low fsc vs. ssc on such a plot which within several
>>seconds
>>equilibrates to a normal appearance of midscale when acquiring normal mouse
>>PBL's.
>>BD tells us and I agree that this is simply a result of the machine
>>equilibrating pressure
>>and that most wait a few seconds before acquiring after placing the sample
>>on the
>>intake port.  In fact this also relates to the question that has appeared
>>elsewhere on this list concerning
>>carryover from sample to sample.  It is clear that this fact was not
>>apparent on older machines
>>because the equilibration had resolved itself before the slower processors
>>could pick it up.
>>Thus my question is, was it necessary to wait a few seconds on the older
>>machines anyway
>>because the finalization of the 'data-values' was commited to by the
>>cytometer or wasn't it
>>necessary because only now with the more rapid CPU does it become truly
>>necessary to
>>wait a few seconds for equilibration prior to acquisition?
>>
>>Gene/UCONN Health
>
>Hi,
>
>I've only seen this when their was a pressure problem on my FACSCalibur.
>When I changed the hydrophobic filters in the side panel and cleaned out
>the cuvette the effect went away.
>
>I used to have the old HP computers running on my FACScan. these computers
>had Direct Memory Access (DMA) that allowed the data generated on the
>FACScan to be dumped directly into the HP's memory. So what you saw on
>screen was real time acquisition of your data. Processor speed is not
>necessarily the bottleneck when it comes to data acquisition. I had a
>Quadra attached to my FACSCalibur before I upgraded to a G3, and saw a
>delay in acquisition. There must have been some buffering going on. The
>NuBUS interface in the Quadra operated at 20MHz but only transferred data
>to memory at 10MHz. Even at these speeds the memory should be adequately
>updated to allow real time acquisition to be displayed. There must have
>been some other bottleneck that meant buffering was needed, does anyone
>know?
>
>In answer to your question, I do not believe that the effect that you are
>seeing is normal and I would ask more questions to BD about it. On the
>HP/FACScan system you did not have to wait and the data was always
>available immediately. As for the G3/FACSCalibur, again I see the data
>available immediately. However I do agree that the pressure system need
>time to equilibrate, but I think a few seconds is too long for your
>instrument, I am guessing but I would say a few milli seconds would be
>more appropriate. I would recheck the tubing, filters and the cuvette for
>blockages.
>
>
>Hope this helps.
>
>
>Andy.
>
>