Actually, 5 cells can mean something in an analysis, although one has to be
cautious. If you are just setting quadrants or gates based upon isotype
controls, then it is not meaningful. With distictly abnormal tumor cells,
you can have a pattern of staining that is far removed from background.
Furthermore there is a cluster, not a spread, and this cluster is seen in
multiple parameter analysis, including size and side scatter. I most often
seen this with Hairy cell leukemia.If 1)the cells are clustering in an
abnormally bright pattern for CD20 and CD22 and positive for kappa but
negative for lambda, 2) lying between normal lymphs and monocytes on FSC
vrs SSC 3) and others tubes are consistent-e.g. Bright CD25+, Bright
CD11c+, Bright CD20 and CD103+ population that forms a nice cluster,this is
detection of tumor in a patient with a history of hairy cell leukemia
matching this immunophenotype. It is the homogeneous, tight clustering
pattern over multiple parameters that allows one to see that this is not
background. This is not generally applicable, but does apply in some
clinical specimens. I still would cautiously word a written diagnosis on
such a case but would verbally convey my impression to the clinician. Now
this may not pertain to the discussion at hand, but I felt chatty this
morning.

	Maryalice

>Ann - -I am a little puzzled. The statistics of a 0.05% population in 10,000
>total cells is only 5 cells. Even with the best possible background, the
>confidence limits of that frequency are statistically pretty poor. In the
>paper I gave to Hai Qi (see corrected reference on Flow Cytometry Mailing
>List), we explored this issue and showed that accurate frequency prediction
>is relatively poor below about 20 cells and only really gets stable at about
>50 cells. If you see something that I don't, I would really want to know!
>--Jim Leary
>
>-----Original Message-----
>From: Ann Atzberger [mailto:Ann.Atzberger@EMBL-Heidelberg.de]
>Sent: Friday, December 03, 1999 7:32 AM
>To: cyto-inbox
>Subject: Re: rare event
>
>
>
>Hallo Hai Qi,
>
>if I understand correctly, you ran a total of 200 thousand cells of the
>same sample through the machine twice. As you say the subpopulation is not
>that rare at 0.15%.  I have often analysed samples with rare events at
>0.05% of the total population with good precision,(granted the signal was
>way above background) and only saving 10,000 events.
>
>If you have a lot of debris or dead cells in the sample the %s can vary; so
>you can either: 1.use a ficoll gradient to clean up the sample.
>2.reset your threshold value to cut out as much debris as possible.
>3.live gate.
>4.Run the cells slowly through the machine.
>5.Make sure the machine is clean, as bits of junk stuck in the sample probe
>can effect your values.
>
>I might be wrong but I think saving 200,000 cells has to be more than ample
>to detect this subpopulation.
>
>Ann
>At 16:49 01.12.99 -0600, you wrote:
>>
>>
>>It is really not too rare, a cell population about 0.04% to 0.15%, while
>the
>>background is 0.01-0.05%. The problem is the variation among different
>>measurements, presumably due to the systematic error. For example, I had
>>10e7 total cells and ran 2*10e5 out of them for two times: one time I got
>>0.04% and the other time I got 0.15%. In order to have smaller variation,
>>obviously I have to increase my sample size, but the last thing I want to
>do
>>is to run through all 10e7 cells. Can anyone give me a handy statistical
>>method for estimating the sample size that makes a cost-effective
>>compromise? Thank you very much.
>>
>>Hai Qi
>>Dept. of Pathology, UTMB
>>
>>