Re: 4 colours on FACS Calibur

From: Karenrt@AOL.COM
Date: Mon Nov 29 1999 - 17:28:13 EST


There is another alternative to PerCP for 4-color analysis on the 
FACSCalibur. PE-Cy7 is available from Caltag Laboratories and was formulated 
specifically to get around this problem. The diode laser on the FACSCalibur 
does excite Cy5, causing compensation problems when using it with APC. Cy7, 
however, is not excited by the red diode. There is very little if any 
compensation between FL3 and FL4 with this conjugate, and about 30% between 
FL2 and FL3, For a list of antibodies conjugated to this fluorochrome, check 
our Website at www.caltag.com.

Karen R. Tamul, MS, MT(ASCP)SI
Mid-Atlantic Territory Manager
Caltag Laboratories Inc.