Re: Texas Red Excitation and Emission

From: Marty Bigos (bigos@stanford.edu)
Date: Wed Nov 03 1999 - 14:04:04 EST


Richard-

I have enclosed excitation and emission spectra of TR as PICT and 
TIFF files; hopefully you can open one of them.

At 514 nm the excitation efficiency of TR is only a few %. Thus it 
will be very difficult to see much antibody expression with this 
choice of fluorochrome.  We use a dye laser set to 595 nm for TR 
measurements (this has been done here for over 10 years). At that 
excitation wavelength TR is a reasonably good fluorochrome.

Hopefully your user can be educated to choose a more suitable fluorochrome.



>I have an investigator that has designed an experiment that lends itself to
>using Texas Red as the fluorochrome of choice.  She has acquired an antibody
>congugated with Texas Red for an assay that is specific to her needs.
>Unfortunately, she failed to inform the operator of our Facstar Plus about
>her choice of fluorochrome.  Our instrument is equipped with a dual laser
>system with one laser limited to a single 488nm line.  The second laser has
>multi-line capabilities, but can only be tuned to the maximum of 514nm.  The
>optimal excitation wavelength for Texas Red as I have discovered is 590nm.
>I have discussed this with her, but she insists on analysis at 514nm.  I
>cannot envision getting any suitable results.  Does anyone out there have
>any experience with Texas Red that could shed any light on this problem?
>
>Thank You
>
>Richard J. Jaramillo
>LRRI
>Albuquerque,NM
>Ph(505)845-1217  Fax(505)845-1198
>rjaramil@lrri.org




Marty Bigos Operations Manager Stanford Shared FACS Facility bigos@stanford.edu (voice) 650-723-6959 (fax) 650-725-8564