Dear Nancy I have had good results using DAKO's conjugated CD30 antibody on the rare cases of anaplastic large cell lymphomas that I've encountered. Normal surface staining using established whole blood methods or usual procedures for handling FNA or lymph node tissues proved no problem. As an adjunct, I would be wary of refering to anaplastic lymphomas as being of T cell orign! Regards Len Brown Brisbane, Oz Nancy Gin wrote: > > Hello, > Would like to know if any of you have successfully > identify Anaplastic T-Cell Lymphoma by using CD30 by > flow cytometry. Any information on manufacturer/clone, > procedure used would be very much appreciated. Thanks > in advance. > Nancy Gin - United Hospital, St. Paul, MN > > ===== > > __________________________________________________ > Do You Yahoo!? > Bid and sell for free at http://auctions.yahoo.com