Re: PE, PE-Cy5, PE-Cy7?

From: Marty Bigos (bigos@stanford.edu)
Date: Tue Sep 28 1999 - 12:33:28 EST


Hi-

We have run these 3 dyes in our facility for quite a while. There are
several technical issues involved.

The first is to make sure that the two tandam conjugates have good energy
transfer. This can vary from lot to lot.

Second, for the Cy7PE especially, but for the others as well, you want to
make sure that you have the optimum light collection setup.  This means
that a) you need red-sensitive pmt's (hamamatsu R3896 or equivalent) and b)
optimal filters. With those 3 dyes a 575/25 bandpass for PE, 700/80
bandpass for Cy5PE, and a 750 long pass for Cy7PE are good choices. Also
the beam splitters need to match the above filter set-up as well.

The reason for all the above hulla-baloo is that when you are collecting
out in the far red, the actual amount of light there is relatively low,
much lower that for FITC, even though the separation between the unstained
background and stained cells may be quite good. So although single stains
with each reagent may look fine, when doing multiple stains, the
compensation interactions and the inherant statistical variation in the low
level light signals can cause problems (smearing of the backgrounds,
mainly). Below is the spillover values for these three colors as well as
FITC.
		Fluor	PE	Cy5PE	Cy7PE
	Fluor	1.0000	0.3532	0.0712	0.0044
	PE	0.0091	1.0000	0.2588	0.0150
Source	Cy5PE	0.0004	0.0241	1.0000	0.0549
	Cy7PE	0.0043	0.2138	0.0822	1.0000
For all the PE-based reagents, there are enough interactions between them
so that full 3x3 compensation is usually necessary for obtaining
interpretable results. The addition of FITC further complicates the
interactions. In general, after compensation you will find that the the
background on all three measurements will increase.

Best of luck.





>Dear All,
>
>does anyone have any experience in analyzing preparations stained with Ab's
>conjugated with PE, PE-Cy5 and PE-Cy7 at the same time. Are there any
>problems associated to this kind of stainings?
>
>
>Carl-Magnus Högerkorp
>Stem cell laboratory
>Department of Internal Medicine
>University Hospital of Lund
>Sweden
Marty Bigos
Operations Manager
Stanford Shared FACS Facility
bigos@stanford.edu
(voice) 650-723-6959
(fax) 650-725-8564



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:54:01 EST