Alessandra Nardin wrote:

> 
> Dear all:
> 
> we are searching for a dead-cell exclusion die to be detected in FL-4 (we
> have a 2 laser FACSCalibur). Has any of you had any experience with TO-PRO-3
> (sold by Molecular Probes), so to either suggest or warmly disencourage its
> use? Alternatively, did you come across to some other possible candidate?
> Thanks.

You didn't mention the cell type that you were using, however we've 
had some good results with this dye with bacteria on an Elite 
(633ex / 675 em) and on a Microcyte (635ex / >650em). The 
important things were to keep incubation times short (~1 min) and 
concentration low (0.1 uM) and of course appropiate controls must 
be run. If you are using animal cells the timings / concentrations 
etc may be very different. If you are using plant cells you probably 
won't get a good topro3 signal because of the autofluorescence. 

Hope that helps,

Haze
----------------------------------------------------------