Ron, we used to use a bi-Anti-PE system: Stain with PE-antibody. Wash, then stain with bi-anti-PE. Wash, then come in with Av-PE. The last two steps can be repeated ad nauseum (until you are naseous). The anti-PE came from Peter Lansdorp (at the Terry Fox in Vancouver). It amplifies very nicely, but is relatively time-consuming. mr >I am doing two color immunofluorescence, and have an antibody that is PE >conjugated. PE is not bright enough to visualize alone, so I have to >amplify with a secondary. However, I cannot use a goat-anti-mouse, because >I am using that on the first antibody. Is there a source for HRP-anti-PE >that can be used as a secondary? Less desirable, is HRP-anti FITC. > >Thanks > > ron > > >Ronald L. Rabin, M.D. >Clinical Associate >Cytokine Biology Unit, Laboratory of Clinical Investigation >National Institute of Allergy and Infectious Diseases >National Institutes of Health >Bldg. 10/Rm. 11N228 >10 Center Drive MSC 1888 >Bethesda, MD 20892-1888 >Phone: (301) 402-4910 >FAX: (301) 402-0627 >