Re: Drosophila discs and cell cylce analysis

From: R.Wadley (s9803537@pop3.unsw.edu.au)
Date: Sun Aug 29 1999 - 23:22:51 EST


	Dear Andrew,

	I have done GFP & cell cycle work, but not with drosophila disks.  I
suggest using Hoerchst for the following reasons.
1/	If your GFP is not bound to organelles or structures within the cell it
will be lost when you puncture the cell wall to let the PI in.  I have seen
this using a Laser Scanning Cytometer, I was able to read the GFP signal,
then stain the same cells with PI & overlay the 2 results.  Virtually all
the GFP was lost in the few minutes it took to perform the PI staining.
2/	GFP can be sensitive to pH changes within the cell, ie its fluorescence
can be reduced or quenched by cell death or some fixation techniques.  I
have not seen this in relation to cell cycle specifically, but have seen it
in sorted samples.
	  
	Hope this helps.

	Regards
	
	Rob W.

At 14:03 08/26/1999 +0100, you wrote:
>Hi All,
>We have a researcher that would like to look at GFP and cell cycle in 
>drosophila disks using PI or Hoechst.  Does anyone out there have a 
>labeling protocol for cells containing GFP for cell cycle?
>Thanks,
>Andy Morschauser
>University of Pennsylvania Cancer Center
>Flow Cytometry & Cell Sorter Shared Facility


R. Wadley, B.App.Sc. M.L.S, Grad.Dip.Sc.MM
Laboratory Manager
Cellular Analysis Facility
School of Microbiology & Immunology
UNSW, New South Wales, Australia, 2052
Ph (BH) 	+61 (2) 9385 3517
Ph (AH)	+61 (2) 9555 1239
Fax 	+61 (2) 9385 1591
E-mail	r.wadley@unsw.edu.au
www	http://www.micro.unsw.edu.au/caf.html