Re: Cy7-APC on 1991 Elite?

From: Howard Shapiro (hms@shapirolab.com)
Date: Fri Jul 02 1999 - 16:40:12 EST


Jim Weaver asks:
>
>Does anyone know if the PMTs in a 1991 vintage Coulter Elite have
>sufficient far red sensitivity to 'see' antibodies labeled with Cy7-APC
>(PharRed et al) conjugates? Thanks for you help.
>

Those tubes are probably R928's; if so, while they probably don't have the
sensitivity, they can be unplugged and replaced with R3896's, which do - for
around $1200 each if you buy the tube quantity one direct from Hamamatsu.

-Howard