Most current commercial cytometers tail off at 500nm scatter detection. trigger should be side scatter, preferential peak signal. Initial strategy is to use 500nm yellow/green beads from polysciences. Dilute particles to reach approx. 10^6/ml (see formula in their particle catalog). Look at green fluorescence versus sisdescatter. Threshold on green fluorescence with your normal lymphocyte settings.Once you see the fluorescent signals lower threshold down to noise level. Adjust scatter to have beads in the middle of the side scatter scale. Change to side scatter triggering with a setting at mid scale and reduce slowly. You should eventually see non-fluorescent noise signals in the low scatter area. Run filtered sample liquid as a control. On the XL we routinely analyse 320 nm particles triggered on peak side scatter, but it requires a clean sheath system. Good luck Gerhard ______________________________ Reply Separator _________________________________ Subject: Microparticle Analysis Using FACScan/Sort Author: laurie_stojanovic@baxter.com at INTERNET Date: 02/04/1999 20:15 Has anyone out there used a Becton Dickinson FACScan or FACSort flow cytometer to analyze platelet- or WBC-derived microparticles (0.2-1.0 um in size)? I am looking for assistance in the following arenas: 1. Distinguishing background (noise) events from "real" microparticle events; looking for ways to increase the limit of sensitivity of these instruments (Becton Dickinson Suggests limit is 0.5 to 1.0 um). 2. Setting used during acquisition; especially, threshold trigger (FSC or FL) and values thereof 3. Gating strategies 4. Sample collection and staining Please respond to laurie_stojanovic@baxter.com with any information you have that could help me. I may also be reached at (847)270-4422. Thanks! Laurie Stojanovic
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:53:22 EST